This thread is for owners of the Little Giant 9300 Incubator to compare the digital controls on there Little Giant 9300, It is not for discussing hatches, hatch results, bird types, ect .
My Little Giant 9300 arrived last week and It looked and worked great, up until I decided I should be running the temp at 100.5 for still air as my local extension office suggests. This is when I discovered what appears to be programming errors within the digital controller unit. The controller unit appears to be right on the money when it comes to reading temperature from it's sensor bulb. However the issue became apparent when setting a temperature other that 99.5 degrees. Since I have eggs in the incubator I am only will do my testing in a range that won't harm the eggs as I'd like them to hatch. We all know the 9300's Humidity meter doesn't read correctly, so we don't need to discuss that.
These are my results.
Setting heater(on) heater(0ff) Maintains
---------------------------------------------------------
99.5 99.5 100.0 99.5
100.0 99.0 99.5 99.5
100.5 99.5 100.0 100.0
101.0 100.0 100.5 101.5
101.5 100.5 101.0 101.0
It could be I just have a bad controller unit or maybe there's more to it than that.
Bottom line: The heater should not shutoff until the control unit display reads the set temperature and should go on below the set temperature.
Anybody else do any testing of their control unit and get these results ?
My Little Giant 9300 arrived last week and It looked and worked great, up until I decided I should be running the temp at 100.5 for still air as my local extension office suggests. This is when I discovered what appears to be programming errors within the digital controller unit. The controller unit appears to be right on the money when it comes to reading temperature from it's sensor bulb. However the issue became apparent when setting a temperature other that 99.5 degrees. Since I have eggs in the incubator I am only will do my testing in a range that won't harm the eggs as I'd like them to hatch. We all know the 9300's Humidity meter doesn't read correctly, so we don't need to discuss that.
These are my results.
Setting heater(on) heater(0ff) Maintains
---------------------------------------------------------
99.5 99.5 100.0 99.5
100.0 99.0 99.5 99.5
100.5 99.5 100.0 100.0
101.0 100.0 100.5 101.5
101.5 100.5 101.0 101.0
It could be I just have a bad controller unit or maybe there's more to it than that.
Bottom line: The heater should not shutoff until the control unit display reads the set temperature and should go on below the set temperature.
Anybody else do any testing of their control unit and get these results ?